I have upgrade XGPRO software today ( to version 12.01 ) after that i received a request to upgarde also Firmware of the TL866II PLUS
during the reflashing the indicators went very slow... (on last 2 years this process was very fast)
after flashing there was a message like "can't restart programmer"

and now i have this issue on "check device id"

there is a way to FORCE a new reflash (maybe using another USB prot with another chipset)
作者: admin    时间: 2022-6-18 12:23

回复 1# Bartoloni
reflash it. donot use external USB hub.  or do it on another computer.
作者: Bartoloni    时间: 2022-6-18 15:40

回复  Bartoloni
reflash it.

The "reflash firmware" is grayed out beacuse the firmware is the latest one ( 04.02.129 ) .. i need to to overwrite it with the same firmware (and i will use a different computer with a different USB chipset)

there is some sort of extra command to force the re-flash??? something like: xgpro.exe /force  ???
作者: Bartoloni    时间: 2022-6-20 18:37

i need to wait next official firmware upgrade ?

meanwhile i bought another TL866II Plus ... i think this device is very useful and i need it...
there is a way to duplicate the firmware from my new device to the old one?
作者: Evcon    时间: 2022-6-20 20:52

Can you take a screenshot of the program after executing the Help-About command on the failed 866?
作者: Bartoloni    时间: 2022-6-20 23:02

sure :)
here you are:


the system self-check is also reporting everything NORMAL.
作者: Evcon    时间: 2022-6-21 02:48

回复 6# Bartoloni


   Then in what case do you get the error? Can you upload a screenshot of the error?
作者: Bartoloni    时间: 2022-6-21 18:40

Can you upload a screenshot

Sure:

作者: Albano.Matos    时间: 2022-6-21 21:42

Sure:
Bartoloni 发表于 2022-6-21 18:40
Open full screen. At bottom, you have a tap to disable ID check. It helps?
Post images
作者: Evcon    时间: 2022-6-21 21:44

This programmer has a forced firmware programming mode. You need to ask the forum administrator for help.
作者: Bartoloni    时间: 2022-6-21 23:34

回复 10# Evcon

:( i think i will do a firmware reflash using the DavidGriffith opensources utilities... but i'm little worried about the bootloader or losing  the serial number...  ican't wait 6 months to reflash with next official update :(
作者: admin    时间: 2022-6-22 14:11

回复 11# Bartoloni

Open the cover of  the programmer .there is a 6P connector,  short PIN2 and PIN5.  then plug into USB.
you can reflash now!
作者: Bartoloni    时间: 2022-6-22 19:49

Very thank you!
作者: Bartoloni    时间: 2022-6-22 20:17

is this the correct pinout?



nevermind... yep, just tried, and the squared one is the 1 :) thank you again for the support!